Question 4
A force sensor can require a zero correction
A uniaxial sensor reads 0.4 divisions at zero force and 2.4 divisions at a known positive 10 N load. Assume a linear calibration. A new reading is 1.6 divisions.
Diagram shown in the original SVG worksheet.
Find force as a function of reading R.
Infer the new load.
Why is multiplying R directly by 5 N per division incorrect?
Question 4 - Solution
A force sensor can require a zero correction
Diagram shown in the original SVG worksheet.
(a) Use the change from zero.
The slope is N per division, giving N for numerical R in divisions.
(b) Apply the offset.
At R=1.6, N.
(c) Keep the intercept.
Using gives 8 N and predicts a nonzero 2 N load at the zero-force reading. A calibration slope alone is insufficient when the sensor has an offset.