Measuring Temperature — Question 3

Section 18-3 · Physics pages 2856–2857

Question 3 A quadratic resistance calibration needs its full inverse A resistance sensor obeys R/R0 = 1 + a∆T + b(∆T)2, where R0 = 100 Ω, a = .0040 K−1, b = 1.0 × 10−5 K−2, and ∆T = T −300 K. Its validated domain is 0 ≤∆T ≤100 K. A bath gives R = 121.0 Ω. T R A calibrated curved response; the local tangent controls small-error sensitivity. (a) Find the physical temperature and the error from ignoring the quadratic term. (b) At this operating point find dR/dT and estimate the temperature half-width from a small resistance half-width .10 Ω. Label the result as a local linear estimate.
Show solutionHide solution
Question 3 - Solution A quadratic resistance calibration needs its full inverse T R (a) Method and calculation 10−5(∆T)2 + .004∆T −.21 = 0. ∆T = −.004 + √ .0042 + 4(10−5)(.21) 2 × 10−5 K ≈46.982 K. Thus T = 346.982 K. The other root is outside the domain. A linear inverse gives 52.5 K, about 5.518 K too high. (b) Interpretation and further calculation dR dT = R0(a + 2b∆T) ≈.49396 Ω/K. δT ≃ .10 .49396 K = .20244 K. This sensitivity estimate is not a substitute for exact endpoint inversion when a strict hard interval is required. Physical check The derivative is positive throughout the validated domain, ensuring a unique inverse there.

Question 3

A quadratic resistance calibration needs its full inverse

A resistance sensor obeys R/R0=1+aΔT+b(ΔT)2R/R_0=1+a\Delta T+b(\Delta T)^2, where R0=100ΩR_0=100\,\Omega, a=.0040K−1a=.0040\,\mathrm{K^{-1}}, b=1.0×10−5K−2b=1.0\times10^{-5}\,\mathrm{K^{-2}}, and ΔT=T−300K\Delta T=T-300\,\mathrm K. Its validated domain is 0≤ΔT≤100K0\le\Delta T\le100\,\mathrm K. A bath gives R=121.0ΩR=121.0\,\Omega.

Diagram shown in the original SVG worksheet.

  1. Find the physical temperature and the error from ignoring the quadratic term.

  2. At this operating point find dR/dTdR/dT and estimate the temperature half-width from a small resistance half-width .10Ω.10\,\Omega. Label the result as a local linear estimate.

Question 3 - Solution

A quadratic resistance calibration needs its full inverse

Diagram shown in the original SVG worksheet.

(a) Method and calculation

10−5(ΔT)2+.004ΔT−.21=0.10^{-5}(\Delta T)^2+.004\Delta T-.21=0.ΔT=−.004+.0042+4(10−5)(.21)2×10−5K≈46.982K.\Delta T=\frac{-.004+\sqrt{.004^2+4(10^{-5})(.21)}}{2\times10^{-5}}\,\mathrm K\approx46.982\,\mathrm K.Thus T=346.982KT=346.982\,\mathrm K. The other root is outside the domain. A linear inverse gives 52.552.5 K, about 5.5185.518 K too high.

(b) Interpretation and further calculation

dRdT=R0(a+2bΔT)≈.49396Ω/K.\frac{dR}{dT}=R_0(a+2b\Delta T)\approx.49396\,\Omega/\mathrm K.δT≃.10.49396K=.20244K.\delta T\simeq\frac{.10}{.49396}\,\mathrm K=.20244\,\mathrm K.This sensitivity estimate is not a substitute for exact endpoint inversion when a strict hard interval is required.

Physical check

The derivative is positive throughout the validated domain, ensuring a unique inverse there.