A blank distinguishes sample signal from background
A color test gives a response of 0.18 unit for water plus reagent and 0.53 unit for a sample plus the same reagent. The instrument reads zero before either mixture is placed in it. Assume additive, stable background.
Diagram shown in the original SVG worksheet.
Explain why the water–reagent mixture is a useful blank.
Calculate the sample-associated response after blank correction.
Name one reason the correction could fail.
Question 4 - Solution
A blank distinguishes sample signal from background
Diagram shown in the original SVG worksheet.
(a) Identify background
The blank includes contributions from reagent, solvent and measurement vessel under the same procedure. Instrument zero alone does not necessarily remove these contributions.
(b) Subtract matched background
The result is in response units, not automatically a concentration.
(c) Check assumptions
The correction can fail if reagent amounts differ, the blank changes with time, or sample components alter the background nonadditively. A valid blank must represent the background present in the actual sample measurement.
Check and interpretation
A corrected response still needs a calibration relation before it can be converted into composition.