Measuring Radiation Dosage — Question 9

Section 42-7 · Physics pages 7128–7129

Question 9 Ionization charge measures deposited energy through the mean pair energy An ideal gas ionization chamber collects charge Qc = 1.60 nC without recombination. Its gas mass is mg = 10.0 mg. Mean energy per ion pair is W=34.0 eV. Use e = 1.60 × 10−19 C and 1 eV = 1.60 × 10−19 J. gas collected charge Qc Schematic; labels, not drawing lengths, specify the data. (a) Find the number of ion pairs and the energy deposited in the gas. (b) Find the gas absorbed dose and explain why it does not directly equal a tissue dose without further calibration.
Show solutionHide solution
Question 9 - Solution Ionization charge measures deposited energy through the mean pair energy gas collected charge Qc (a) Method and calculation Npairs = Qc/e = 1010 . Eg = NpairsW = 1010(34.0)(1.60 × 10−19) = 5.44 × 10−8 J . (b) Interpretation and further calculation Dg = Eg/mg = 5.44 × 10−8/10−5 = .00544 Gy = 5.44 mGy . Dtissue̸ = Dg without a material-response conversion . Physical check Collected charge counts one elementary charge per ion pair in the external circuit, not twice that charge. W is a mean expenditure per pair, larger than a single ionization threshold.

Question 9

Ionization charge measures deposited energy through the mean pair energy

An ideal gas ionization chamber collects charge Qc=1.60nCQ_c=1.60\,\mathrm{nC} without recombination. Its gas mass is mg=10.0mgm_g=10.0\,\mathrm{mg}. Mean energy per ion pair is W=34.0 eV. Use e=1.60×10−19Ce=1.60\times10^{-19}\,\mathrm C and 1eV=1.60×10−19J1\,\mathrm{eV}=1.60\times10^{-19}\,\mathrm J.

Diagram shown in the original SVG worksheet.

  1. Find the number of ion pairs and the energy deposited in the gas.

  2. Find the gas absorbed dose and explain why it does not directly equal a tissue dose without further calibration.

Question 9 - Solution

Ionization charge measures deposited energy through the mean pair energy

Diagram shown in the original SVG worksheet.

(a) Method and calculation

Npairs=Qc/e=1010.\boxed{N_{\mathrm{pairs}}=Q_c/e=10^{10}}.Eg=NpairsW=1010(34.0)(1.60×10−19)=5.44×10−8J.\boxed{E_g=N_{\mathrm{pairs}}W=10^{10}(34.0)(1.60\times10^{-19})=5.44\times10^{-8}\,\mathrm J}.

(b) Interpretation and further calculation

Dg=Eg/mg=5.44×10−8/10−5=.00544Gy=5.44mGy.\boxed{D_g=E_g/m_g=5.44\times10^{-8}/10^{-5}=.00544\,\mathrm{Gy}=5.44\,\mathrm{mGy}}.Dtissue≠Dg without a material-response conversion.\boxed{D_{\mathrm{tissue}}\ne D_g\text{ without a material-response conversion}}.

Physical check

Collected charge counts one elementary charge per ion pair in the external circuit, not twice that charge. W is a mean expenditure per pair, larger than a single ionization threshold.