The Ammeter and the Voltmeter — Question 8

Section 27-7 · Physics pages 4446–4447

Question 8 A higher voltmeter range trades reduced loading against a larger absolute specification An analog voltmeter has input resistance 10.0 kilohm per volt of full-scale range and a maximum indication error of .5 percent of full scale. A source has E=10.0 V and r=20.0 kilohm. Compare its 10.0 V and 100 V ranges. + − E r V Circuit connections and marked polarities define the model; wire lengths are schematic. (a) Find the loaded voltage on each range. (b) Bound the worst absolute difference between indicated voltage and E by adding loading error and the full-scale error allowance. Which range has the smaller bound?
Show solutionHide solution
Question 8 - Solution A higher voltmeter range trades reduced loading against a larger absolute specification + − E r V (a) Method and calculation R10 = 100 kΩ, R100 = 1.00 MΩ. V10 = 10100 120 = 25/3 V, V100 = 101000 1020 = 500/51 V. (b) Interpretation and further calculation B10 = 10 −25/3 + .005(10) = 103/60 V ≃1.72 V. B100 = 10 −500/51 + .005(100) = 71/102 V ≃.696 V. The 100 V range has the smaller specified worst-case bound here. Physical check Range choice depends on both circuit loading and the instrument error model; nominal resolution alone is insufficient.

Question 8

A higher voltmeter range trades reduced loading against a larger absolute specification

An analog voltmeter has input resistance 10.0 kilohm per volt of full-scale range and a maximum indication error of .5 percent of full scale. A source has E=10.0 V and r=20.0 kilohm. Compare its 10.0 V and 100 V ranges.

Diagram shown in the original SVG worksheet.

  1. Find the loaded voltage on each range.

  2. Bound the worst absolute difference between indicated voltage and E by adding loading error and the full-scale error allowance. Which range has the smaller bound?

Question 8 - Solution

A higher voltmeter range trades reduced loading against a larger absolute specification

Diagram shown in the original SVG worksheet.

(a) Method and calculation

R10=100kΩ,R100=1.00MΩ.R_{10}=100\,\mathrm{k\Omega},\quad R_{100}=1.00\,\mathrm{M\Omega}.V10=10100120=25/3V,V100=1010001020=500/51V.V_{10}=10\frac{100}{120}=25/3\,\mathrm V,\quad V_{100}=10\frac{1000}{1020}=500/51\,\mathrm V.

(b) Interpretation and further calculation

B10=10−25/3+.005(10)=103/60V≃1.72V.B_{10}=10-25/3+.005(10)=103/60\,\mathrm V\simeq1.72\,\mathrm V.B100=10−500/51+.005(100)=71/102V≃.696V.B_{100}=10-500/51+.005(100)=71/102\,\mathrm V\simeq.696\,\mathrm V.The 100 V range has the smaller specified worst-case bound here.

Physical check

Range choice depends on both circuit loading and the instrument error model; nominal resolution alone is insufficient.