Measuring Temperature — Question 9

Section 18-3 · Physics pages 2868–2869

Question 9 Digit size competes with calibration uncertainty Thermometer A has slope .20 V/K, rounds voltage to the nearest .002 V, and has an additional independent hard calibration error of .030 K. Thermometer B has slope .050 V/K, rounds to the nearest .0002 V, and has calibration error .010 K. Use closed half-bin bounds and add worst-case contributions. instrument A instrument B resolution plus calibration Schematic; labels, not drawing lengths, specify the data. (a) Find each total temperature half-width. Which instrument meets a requirement of at most .015 K? (b) If A displays ten times more voltage digits without changing its calibration error, can it meet that requirement? Quantify its best resulting bound.
Show solutionHide solution
Question 9 - Solution Digit size competes with calibration uncertainty instrument A instrument B resolution plus calibration (a) Method and calculation δTA = .002/2 .20 + .030 = .035 K. δTB = .0002/2 .050 + .010 = .012 K. Only B meets .015 K despite its smaller voltage sensitivity. (b) Interpretation and further calculation δTA,finer = .0002/2 .20 + .030 = .0305 K. lim digit size→0 δTA = .030 K > .015 K. Calibration dominates, so finer display increments alone cannot meet the specification. Physical check The contributions are stipulated hard bounds; statistical independence does not justify replacing the worst-case sum by a root-sum-square.

Question 9

Digit size competes with calibration uncertainty

Thermometer A has slope .20V/K.20\,\mathrm{V/K}, rounds voltage to the nearest .002V.002\,\mathrm V, and has an additional independent hard calibration error of .030K.030\,\mathrm K. Thermometer B has slope .050V/K.050\,\mathrm{V/K}, rounds to the nearest .0002V.0002\,\mathrm V, and has calibration error .010K.010\,\mathrm K. Use closed half-bin bounds and add worst-case contributions.

Diagram shown in the original SVG worksheet.

  1. Find each total temperature half-width. Which instrument meets a requirement of at most .015K.015\,\mathrm K?

  2. If A displays ten times more voltage digits without changing its calibration error, can it meet that requirement? Quantify its best resulting bound.

Question 9 - Solution

Digit size competes with calibration uncertainty

Diagram shown in the original SVG worksheet.

(a) Method and calculation

δTA=.002/2.20+.030=.035K.\delta T_A=\frac{.002/2}{.20}+.030=.035\,\mathrm K.δTB=.0002/2.050+.010=.012K.\delta T_B=\frac{.0002/2}{.050}+.010=.012\,\mathrm K.Only B meets .015.015 K despite its smaller voltage sensitivity.

(b) Interpretation and further calculation

δTA,finer=.0002/2.20+.030=.0305K.\delta T_{A,\mathrm{finer}}=\frac{.0002/2}{.20}+.030=.0305\,\mathrm K.limdigit size→0δTA=.030K>.015K.\lim_{\text{digit size}\to0}\delta T_A=.030\,\mathrm K>.015\,\mathrm K.Calibration dominates, so finer display increments alone cannot meet the specification.

Physical check

The contributions are stipulated hard bounds; statistical independence does not justify replacing the worst-case sum by a root-sum-square.