Temperature — Question 9

Section 18-1 · Physics pages 2828–2829

Question 9 A finite probe changes the sample it measures A sample initially at unknown Ts contacts a probe initially at 280 K. Their thermal parameters are Cs = 40 J/K and Cp = 10 J/K. The isolated pair reaches Tf = 304 K. Use the supplied balance Cs(Tf −Ts) + Cp(Tf −Tp) = 0, with constant parameters and no phase change. sample Cs probe Cp isolated contact Schematic; labels, not drawing lengths, specify the data. (a) Infer the original sample temperature and the error of calling Tf its original value. (b) For a sample initially at 310 K and a probe at 280 K, determine the largest Cp/Cs that keeps the disturbance no greater than 0.50 K.
Show solutionHide solution
Question 9 - Solution A finite probe changes the sample it measures sample Cs probe Cp isolated contact (a) Method and calculation Ts = Tf + Cp Cs (Tf −Tp) = 304 + 10 40(24) = 310 K. Tf −Ts = −6.0 K. The equilibrium reading belongs to the final combined system. (b) Interpretation and further calculation Let r = Cp/Cs ≥0. Rearranging the supplied balance gives Tf = Ts + rTp 1 + r , Ts −Tf = 30r 1 + r K. 30r 1 + r ≤.50 =⇒r ≤1 59 ≈.01695. A sufficiently small probe reduces disturbance; fast response alone does not guarantee this. Physical check The inferred 310 K exceeds the final 304 K because the initially cooler probe warmed at the sample’s expense.

Question 9

A finite probe changes the sample it measures

A sample initially at unknown TsT_s contacts a probe initially at 280K280\,\mathrm K. Their thermal parameters are Cs=40J/KC_s=40\,\mathrm{J/K} and Cp=10J/KC_p=10\,\mathrm{J/K}. The isolated pair reaches Tf=304KT_f=304\,\mathrm K. Use the supplied balance Cs(Tf−Ts)+Cp(Tf−Tp)=0C_s(T_f-T_s)+C_p(T_f-T_p)=0, with constant parameters and no phase change.

Diagram shown in the original SVG worksheet.

  1. Infer the original sample temperature and the error of calling TfT_f its original value.

  2. For a sample initially at 310K310\,\mathrm K and a probe at 280K280\,\mathrm K, determine the largest Cp/CsC_p/C_s that keeps the disturbance no greater than 0.50K0.50\,\mathrm K.

Question 9 - Solution

A finite probe changes the sample it measures

Diagram shown in the original SVG worksheet.

(a) Method and calculation

Ts=Tf+CpCs(Tf−Tp)=304+1040(24)=310K.T_s=T_f+\frac{C_p}{C_s}(T_f-T_p)=304+\frac{10}{40}(24)=310\,\mathrm K.Tf−Ts=−6.0K.T_f-T_s=-6.0\,\mathrm K.The equilibrium reading belongs to the final combined system.

(b) Interpretation and further calculation

Let r=Cp/Cs≥0r=C_p/C_s\ge0. Rearranging the supplied balance givesTf=Ts+rTp1+r,Ts−Tf=30r1+rK.T_f=\frac{T_s+rT_p}{1+r},\qquad T_s-T_f=\frac{30r}{1+r}\,\mathrm K.30r1+r≤.50⇒r≤159≈.01695.\frac{30r}{1+r}\le.50\ \Longrightarrow\ r\le\frac1{59}\approx.01695.A sufficiently small probe reduces disturbance; fast response alone does not guarantee this.

Physical check

The inferred 310310 K exceeds the final 304304 K because the initially cooler probe warmed at the sample’s expense.