Question 8
Two calibration points reveal both offset and gain
A pressure sensor obeys , where is in volts and gauge pressure is in kilopascals. Against a reference manometer it reads at zero gauge pressure and at .
Diagram shown in the original SVG worksheet.
Infer and , then find the gauge pressure for a reading of .
At a different site atmospheric pressure is . The sensor still reads while referenced to the local atmosphere. Find absolute pressure and explain why the original atmospheric pressure is irrelevant.
Question 8 - Solution
Two calibration points reveal both offset and gain
Diagram shown in the original SVG worksheet.
(a) Derivation and calculation
(b) Result and interpretation
The gauge reference is the atmosphere at its current reference port. Calibration determines gain and electrical offset, not a permanent atmospheric pressure.
Physical check
Using without subtracting would introduce a spurious 2.00 kPa.