Mass Spectrometry of Large Biomolecules — Question 5

Section 15-15 · Organic Chemistry pages 3519–3520

Show solutionHide solution

Question 5

Compare soft ionization strategies for large molecules

An intact biomolecule is too nonvolatile for direct GC analysis.

Diagram shown in the original SVG worksheet.

  1. How can electrospray bring a large molecular mass into a lower m/z window?

  2. Why is MALDI also useful for many biomolecules?

  3. Why must the actual charge state still be verified?

Question 5 - Solution

Compare soft ionization strategies for large molecules

Diagram shown in the original SVG worksheet.

(a)

Multiple charging divides the ion mass by a larger charge number, lowering m/z.

(b)

Matrix-assisted laser desorption/ionization can produce intact ions from many nonvolatile biomolecules under suitable conditions.

(c)

Both methods can produce exceptions, adducts and multiple charge states; a source name alone does not determine every ion assignment.