Question 9
Calibrate an offset pressure sensor
A sensor model is . At a known pressure of 80.0 kPa it reads 82.5 kPa. Later it reads 110.0 kPa. Assume the offset remains constant.
Diagram shown in the original SVG worksheet.
Find the corrected later pressure.
Why does a single calibration not establish that the sensor has the correct scale factor?
Question 9 - Solution
Calibrate an offset pressure sensor
Diagram shown in the original SVG worksheet.
(a) Reasoning and result
kPa. Corrected pressure is kPa.
(b) Analysis
One reference point can determine an offset only after assuming unit slope. A more general model requires at least two independent reference pressures to determine both parameters.
Check and interpretation
The correction is conditional on the specified additive-offset model.