Show solutionHide solution
Question 10
Design a control that tests galvanic coupling
Iron and copper coupons are immersed in the same aerated electrolyte. Compare a metallic wire joining them with an insulating connection. Keep area, electrolyte, temperature and surface preparation the same.
Diagram shown in the original SVG worksheet.
Predict the effect of breaking the electronic connection and state what it does not guarantee.
Propose a measurement and a replicate/control strategy that distinguishes galvanic enhancement from ordinary independent corrosion.
Question 10 - Solution
Design a control that tests galvanic coupling
Diagram shown in the original SVG worksheet.
(a) Reasoning and result
Breaking the connection stops direct galvanic electron exchange between coupons. It does not guarantee zero iron corrosion because local anodic and cathodic sites can still form on iron alone.
(b) Analysis
Measure cleaned iron mass loss over the same exposure time and, in wired trials, coupling current. Use replicated matched coupons and an iron-only control in the same electrolyte; compare losses per area and time.
Check and interpretation
The control changes electrical coupling while holding other conditions as constant as practical.