Uncertainty in Measurement — Question 4

Section 1-4 · Chemistry pages 73–74

Show solutionHide solution

Question 4

Use bracketing checks to detect calibration drift

A stable reference assigned 20.00 g gives balance indications of 20.00, 20.04, 20.08 and 20.12 g at 0, 10, 20 and 30 min. A sample indicates 7.56 g at 15 min. For correction, assume the balance has an additive offset that varies linearly with time and applies equally to the reference and sample.

Diagram shown in the original SVG worksheet.

  1. Find the offset at each reference check and its rate of change.

  2. Estimate the offset at 15 min and the corrected sample mass.

  3. Why does this correction not justify reporting the corrected mass as exact?

Question 4 - Solution

Use bracketing checks to detect calibration drift

Diagram shown in the original SVG worksheet.

(a) Infer an additive offset

Offset equals indication minus assigned reference value. The four offsets are 0.000.00, +0.04+0.04, +0.08+0.08 and +0.12+0.12 g.ΔbΔt=0.12−0.0030−0=0.004g/min.\frac{\Delta b}{\Delta t}=\frac{0.12-0.00}{30-0}=0.004\ \mathrm{g/min}.The repeated equal increments support the assumed linear model over the checks.

(b) Interpolate and correct

b(15)=0.004(15)=0.06g.b(15)=0.004(15)=0.06\ \mathrm g.mcorrected=mindicated−b=7.56−0.06=7.50g.m_{\mathrm{corrected}}=m_{\mathrm{indicated}}-b=7.56-0.06=\boxed{7.50\ \mathrm g}.A positive offset must be subtracted, not added.

(c) Retain residual uncertainty

The model may not describe short-term fluctuations or nonlinear drift between checks. Reading uncertainty, reference uncertainty and any load-dependent response also matter. No numerical residual bound is provided, so an exact corrected value or confidence interval cannot be claimed.

Check and interpretation

The interpolated offset lies between the offsets at 10 and 20 min; the corrected value is lower than the indication.